Thermocouples: temperature to emf (or vice verca)
//The function emf2T converts a thermocouple electromotive force (emf, in milli volts  to temperature (in Celsius) or vice verca;
//currently for B, S and C-Type thermocouples. All coefficients taken from ASTM E-230.
//syntax:               emf2T(TC_string, temp, emf)
//Parameters:       TC_string   is a string indicating the thermocouple type, here B, S or C
//              tmp         is temperature in Celsius
//              emf         is the emf in mV
//                          The value that needs to be calculated must be entered as NaN
//To convert a temperature into an emf value use e.g.   print   emf2T("B", 1400, nan)
//To convert emf to temperature use e.g.            print   emf2T("B", nan, 8.956)
function emf2T(type, Tmp, emf)
    string type
    variable Tmp, emf
    variable c0, c1,c2,c3,c4,c5,c6,c7,c8, c9
    c1 = 0      //if not assigned later make sure they are all zero
    c2 = 0
    c3 = 0
    c4 = 0
    c5 = 0
    c6 = 0
    c7 = 0
    c8 = 0
    c9 = 0
    // ***********************       B- Type thermocouples Pt6%Rh / Pt30%Rh          *************************
    strswitch(type)
        case "B":
        if (numtype(Tmp) == 0 && numtype(emf) == 2)     // this is a temperature to EMF conversion
            if (tmp >= 0 && tmp <= 630.615)
                c0 = 0.000000000000E+00
                c1 =-0.246508183460E-03
                c2 = 0.590404211710E-05
                c3 =-0.132579316360E-08
                c4 = 0.156682919010E-11
                c5 =-0.169445292400E-14
                c6 = 0.629903470940E-18
                c7 = 0
                c8 = 0
                return c0*tmp^0 + c1*tmp^1 +c2*tmp^2 +c3*tmp^3 +c4*tmp^4 +c5*tmp^5 +c6*tmp^6  +c7*tmp^7  +c8*tmp^8
            if (tmp > 630.615 && tmp <= 1820)
                c0 = -0.389381686210E+01
                c1 = 0.285717474700E-01
                c2 =-0.848851047850E-04
                c3 = 0.157852801640E-06
                c4 =-0.168353448640E-09
                c5 = 0.111097940130E-12
                c6 =-0.445154310330E-16
                c7 = 0.989756408210E-20
                c8 =-0.937913302890E-24
        if (numtype(Tmp) == 2 && numtype(emf) == 0)     // this is a EMF to temperature conversion
            if (emf >= 0.291 && emf <= 2.431)
                c0 =     9.8423321E+01
                    c1 =   6.9971500E+02
                c2 =     -8.4765304E+02
                c3 =        1.0052644E+03
                c4 =       -8.3345952E+02
                c5 =       4.5508542E+02
                c6 =      -1.5523037E+02
                c7 =        2.9886750E+01
                c8 =       -2.4742860E+00
                return  c0 + c1*emf^1 +c2*emf^2 +c3*emf^3 +c4*emf^4 +c5*emf^5 +c6*emf^6  +c7*emf^7  +c8*emf^8
            if (emf > 2.431 && emf <= 13.82)
                c0 =    2.1315071E+02
                c1 =    2.8510504E+02
                    c2 =    -5.2742887E+01
                c3 =     9.9160804E+00
                c4 =     -1.2965303E+00
                c5 =      1.1195870E-01
                c6 =     -6.0625199E-03
                c7 =       1.8661696E-04
                c8 =    -2.4878585E-06
        break
        // ***********************       S- Type thermocouples Pt / Pt10%Rh          ******************************
        case "S":
            if (tmp >= -50  && tmp <= 1064.18)
                c1 =0.540313308631E-02
                c2 = 0.125934289740E-04
                c3 =-0.232477968689E-07
                c4 = 0.322028823036E-10
                c5 =-0.331465196389E-13
                c6 = 0.255744251786E-16
                c7 =-0.125068871393E-19
                c8 = 0.271443176145E-23
            if (tmp > 1064.18 && tmp <= 1664.5)
                c0 = 0.132900444085E+01
                c1 =  0.334509311344E-02
                c2 =  0.654805192818E-05
                c3 = -0.164856259209E-08
                c4 =  0.129989605174E-13
            if (tmp > 1664.5  && tmp <= 1768.1)
                c0 =    0.146628232636E+03
                c1 = -0.258430516752E+00
                c2 =  0.163693574641E-03
                c3 = -0.330439046987E-07
                c4 = -0.943223690612E-14
            if (emf >= -0.235 && emf <= 1.874)
                c0  =   0
                c1  =   1.84949460E+02
                c2  =   -8.00504062E+01
                c3  =   1.02237430E+02
                c4  =   -1.52248592E+02
                c5  =   1.88821343E+02
                c6  =   -1.59085941E+02
                c7  =   8.23027880E+01
                c8  =   -2.34181944E+01
                c9  =   2.79786260E+00
                return  c0 + c1*emf^1 +c2*emf^2 +c3*emf^3 +c4*emf^4 +c5*emf^5 +c6*emf^6  +c7*emf^7  +c8*emf^8 +c9*emf^9
            if (emf > 1.874 && emf <=11.950)
                c0  =   1.291507177E+01
                c1  =   1.466298863E+02
                c2  =   -1.534713402E+01
                c3  =   3.145945973
                c4  =   -4.163257839E-01
                c5  =   3.187963771E-02
                c6  =   -1.291637500E-03
                c7  =   2.183475087E-05
                c8  =   -1.447379511E-07
                c9  =   8.211272125E-09
            if (emf > 11.950 && emf <= 17.536)
                c0  =   -8.087801117E+01
                c1  =   1.621573104E+02
                c2  =   -8.536869453E+00
                c3  =   4.719686976E-01
                c4  =   -1.441693666E-02
                c5  =   2.081618890E-04
            if (emf > 17.536 && emf <= 18.693)
                c0  =   5.333875126E+04
                c1  =   -1.235892298E+04
                c2  =   1.092657613E+03
                c3  =   -4.265693686E+01
                c4  =   6.247205420E-01
        break
        // ***********************    C- Type thermocouples W5%Re / W26%Re     *****************************
        case "C":
                c0 = 0.0000000
                c1 = 1.34060323e-2
                c2 = 1.19249923e-5
                c3 = -7.98063543e-9
                c4 = -5.07875153e-12
                c5 = 1.31641973e-14
                c6 = -7.91973323e-18
            if (tmp > 630.615 && tmp <= 2315)
                c0 =  4.05288233e-1
                c1 =  1.15093553e-2
                c2 =  1.56964533e-5
                c3 =  -1.37044123e-8
                c4 =  5.22908733e-12
                c5 =  -9.20827583e-16
                c6 = 4.52451123e-20
            c0 = 0.0
            c1 = 7.4124732e1
            c2 = -4.28082813
            c3 = 5.21138920e-1
            c4 = -4.57487201e-2
            c5 = 2.80578284e-3
            c6 = -1.13145137e-4
            c7 = 2.85489684e-6
            c8 = -4.07643828e-8
            c9 = 2.51358071e-10
        break
        default:
            break
    endswitch
    return NaN   // if non of the other if-statements are entered, return NaN
That's terrific! A lot of effort typing in all those constants!
I would comment that the straightforward method of evaluating a polynomial has potential problems caused by floating-point truncation errors. A better way to do it is outlined in Numerical Recipes. This method is followed by the Igor poly() function, so I would recommend poly(). To use it, you need to create a wave with the coefficients, something like this fragment that implements just one part of your function:
                Make/D/FREE cw = {4.05288233e-1,  1.15093553e-2,   1.56964533e-5, -1.37044123e-8, 5.22908733e-12, -9.20827583e-16, 4.52451123e-20}
                return poly(cw, tmp)
There's a GUI for thermocouple emf calculations here: https://www.wavemetrics.com/node/21411
It does S, B, R, K, T, E, J, N, C, and D types, and includes cold junction compensation. It should convert between emf and temperature in either direction, depending on which setvariable you adjust.
Himmelkreuzdonnerwetter! I've just noticed the identity of the original poster - Hi Christian! I guess there aren't so many people out there using types C and D :) I wonder, was it I who introduced you to the delights of Igor back in Hannover in the late 90s, or did you pick it up somewhere further on down the line?
Hi Tony,
(I haven't been looking here for a while)
yes, indeed! It is all your fault! ;-)
C
Heyy. Anyone know how to apply the cold junction compensation for the thermocouple. Thanks