Thermocouples: temperature to emf (or vice verca)
ChrLie
#pragma rtGlobals=1 // Use modern global access method.
//The function emf2T converts a thermocouple electromotive force (emf, in milli volts to temperature (in Celsius) or vice verca;
//currently for B, S and C-Type thermocouples. All coefficients taken from ASTM E-230.
//syntax: emf2T(TC_string, temp, emf)
//Parameters: TC_string is a string indicating the thermocouple type, here B, S or C
// tmp is temperature in Celsius
// emf is the emf in mV
// The value that needs to be calculated must be entered as NaN
//To convert a temperature into an emf value use e.g. print emf2T("B", 1400, nan)
//To convert emf to temperature use e.g. print emf2T("B", nan, 8.956)
function emf2T(type, Tmp, emf)
string type
variable Tmp, emf
variable c0, c1,c2,c3,c4,c5,c6,c7,c8, c9
c1 = 0 //if not assigned later make sure they are all zero
c2 = 0
c3 = 0
c4 = 0
c5 = 0
c6 = 0
c7 = 0
c8 = 0
c9 = 0
// *********************** B- Type thermocouples Pt6%Rh / Pt30%Rh *************************
strswitch(type)
case "B":
if (numtype(Tmp) == 0 && numtype(emf) == 2) // this is a temperature to EMF conversion
if (tmp >= 0 && tmp <= 630.615)
c0 = 0.000000000000E+00
c1 =-0.246508183460E-03
c2 = 0.590404211710E-05
c3 =-0.132579316360E-08
c4 = 0.156682919010E-11
c5 =-0.169445292400E-14
c6 = 0.629903470940E-18
c7 = 0
c8 = 0
return c0*tmp^0 + c1*tmp^1 +c2*tmp^2 +c3*tmp^3 +c4*tmp^4 +c5*tmp^5 +c6*tmp^6 +c7*tmp^7 +c8*tmp^8
if (tmp > 630.615 && tmp <= 1820)
c0 = -0.389381686210E+01
c1 = 0.285717474700E-01
c2 =-0.848851047850E-04
c3 = 0.157852801640E-06
c4 =-0.168353448640E-09
c5 = 0.111097940130E-12
c6 =-0.445154310330E-16
c7 = 0.989756408210E-20
c8 =-0.937913302890E-24
if (numtype(Tmp) == 2 && numtype(emf) == 0) // this is a EMF to temperature conversion
if (emf >= 0.291 && emf <= 2.431)
c0 = 9.8423321E+01
c1 = 6.9971500E+02
c2 = -8.4765304E+02
c3 = 1.0052644E+03
c4 = -8.3345952E+02
c5 = 4.5508542E+02
c6 = -1.5523037E+02
c7 = 2.9886750E+01
c8 = -2.4742860E+00
return c0 + c1*emf^1 +c2*emf^2 +c3*emf^3 +c4*emf^4 +c5*emf^5 +c6*emf^6 +c7*emf^7 +c8*emf^8
if (emf > 2.431 && emf <= 13.82)
c0 = 2.1315071E+02
c1 = 2.8510504E+02
c2 = -5.2742887E+01
c3 = 9.9160804E+00
c4 = -1.2965303E+00
c5 = 1.1195870E-01
c6 = -6.0625199E-03
c7 = 1.8661696E-04
c8 = -2.4878585E-06
break
// *********************** S- Type thermocouples Pt / Pt10%Rh ******************************
case "S":
if (tmp >= -50 && tmp <= 1064.18)
c1 =0.540313308631E-02
c2 = 0.125934289740E-04
c3 =-0.232477968689E-07
c4 = 0.322028823036E-10
c5 =-0.331465196389E-13
c6 = 0.255744251786E-16
c7 =-0.125068871393E-19
c8 = 0.271443176145E-23
if (tmp > 1064.18 && tmp <= 1664.5)
c0 = 0.132900444085E+01
c1 = 0.334509311344E-02
c2 = 0.654805192818E-05
c3 = -0.164856259209E-08
c4 = 0.129989605174E-13
if (tmp > 1664.5 && tmp <= 1768.1)
c0 = 0.146628232636E+03
c1 = -0.258430516752E+00
c2 = 0.163693574641E-03
c3 = -0.330439046987E-07
c4 = -0.943223690612E-14
if (emf >= -0.235 && emf <= 1.874)
c0 = 0
c1 = 1.84949460E+02
c2 = -8.00504062E+01
c3 = 1.02237430E+02
c4 = -1.52248592E+02
c5 = 1.88821343E+02
c6 = -1.59085941E+02
c7 = 8.23027880E+01
c8 = -2.34181944E+01
c9 = 2.79786260E+00
return c0 + c1*emf^1 +c2*emf^2 +c3*emf^3 +c4*emf^4 +c5*emf^5 +c6*emf^6 +c7*emf^7 +c8*emf^8 +c9*emf^9
if (emf > 1.874 && emf <=11.950)
c0 = 1.291507177E+01
c1 = 1.466298863E+02
c2 = -1.534713402E+01
c3 = 3.145945973
c4 = -4.163257839E-01
c5 = 3.187963771E-02
c6 = -1.291637500E-03
c7 = 2.183475087E-05
c8 = -1.447379511E-07
c9 = 8.211272125E-09
if (emf > 11.950 && emf <= 17.536)
c0 = -8.087801117E+01
c1 = 1.621573104E+02
c2 = -8.536869453E+00
c3 = 4.719686976E-01
c4 = -1.441693666E-02
c5 = 2.081618890E-04
if (emf > 17.536 && emf <= 18.693)
c0 = 5.333875126E+04
c1 = -1.235892298E+04
c2 = 1.092657613E+03
c3 = -4.265693686E+01
c4 = 6.247205420E-01
break
// *********************** C- Type thermocouples W5%Re / W26%Re *****************************
case "C":
c0 = 0.0000000
c1 = 1.34060323e-2
c2 = 1.19249923e-5
c3 = -7.98063543e-9
c4 = -5.07875153e-12
c5 = 1.31641973e-14
c6 = -7.91973323e-18
if (tmp > 630.615 && tmp <= 2315)
c0 = 4.05288233e-1
c1 = 1.15093553e-2
c2 = 1.56964533e-5
c3 = -1.37044123e-8
c4 = 5.22908733e-12
c5 = -9.20827583e-16
c6 = 4.52451123e-20
c0 = 0.0
c1 = 7.4124732e1
c2 = -4.28082813
c3 = 5.21138920e-1
c4 = -4.57487201e-2
c5 = 2.80578284e-3
c6 = -1.13145137e-4
c7 = 2.85489684e-6
c8 = -4.07643828e-8
c9 = 2.51358071e-10
break
default:
break
endswitch
return NaN // if non of the other if-statements are entered, return NaN
string type
variable Tmp, emf
c2 = 0
c3 = 0
c4 = 0
c5 = 0
c6 = 0
c7 = 0
c8 = 0
c9 = 0
strswitch(type)
case "B":
c7 = 0
c8 = 0
c0 = 9.8423321E+01
c1 = 6.9971500E+02
c2 = -8.4765304E+02
c3 = 1.0052644E+03
c4 = -8.3345952E+02
c5 = 4.5508542E+02
c6 = -1.5523037E+02
c7 = 2.9886750E+01
c8 = -2.4742860E+00
c0 = 2.1315071E+02
c1 = 2.8510504E+02
c2 = -5.2742887E+01
c3 = 9.9160804E+00
c4 = -1.2965303E+00
c5 = 1.1195870E-01
c6 = -6.0625199E-03
c7 = 1.8661696E-04
c8 = -2.4878585E-06
break
case "S":
c0 = 0
c1 = 1.84949460E+02
c3 = 1.02237430E+02
c5 = 1.88821343E+02
c7 = 8.23027880E+01
c9 = 2.79786260E+00
c3 = 3.145945973
break
case "C":
c0 = 0.0000000
c1 = 1.34060323e-2
c2 = 1.19249923e-5
c3 = -7.98063543e-9
c5 = 1.31641973e-14
c0 = 4.05288233e-1
c1 = 1.15093553e-2
c2 = 1.56964533e-5
c3 = -1.37044123e-8
c4 = 5.22908733e-12
c6 = 4.52451123e-20
c0 = 0.0
c1 = 7.4124732e1
c2 = -4.28082813
c3 = 5.21138920e-1
c4 = -4.57487201e-2
c5 = 2.80578284e-3
c6 = -1.13145137e-4
c7 = 2.85489684e-6
c8 = -4.07643828e-8
c9 = 2.51358071e-10
break
default:
break
endswitch
I would comment that the straightforward method of evaluating a polynomial has potential problems caused by floating-point truncation errors. A better way to do it is outlined in Numerical Recipes. This method is followed by the Igor poly() function, so I would recommend poly(). To use it, you need to create a wave with the coefficients, something like this fragment that implements just one part of your function:
Make/D/FREE cw = {4.05288233e-1, 1.15093553e-2, 1.56964533e-5, -1.37044123e-8, 5.22908733e-12, -9.20827583e-16, 4.52451123e-20}
return poly(cw, tmp)
June 1, 2012 at 08:55 am - Permalink
There's a GUI for thermocouple emf calculations here: https://www.wavemetrics.com/node/21411
It does S, B, R, K, T, E, J, N, C, and D types, and includes cold junction compensation. It should convert between emf and temperature in either direction, depending on which setvariable you adjust.
June 27, 2012 at 09:43 am - Permalink
June 29, 2012 at 11:17 am - Permalink
(I haven't been looking here for a while)
yes, indeed! It is all your fault! ;-)
C
July 5, 2012 at 03:06 am - Permalink
August 21, 2012 at 02:31 pm - Permalink
After many years, I have released my thermocouple calulator GUI as an IgorExchange project. I'll update the link in this thread, in case anyone out there is interested in such a thing.
September 24, 2020 at 02:26 am - Permalink
